50 Accuracy of operating data
IEC 60601-1:1988,Clause 50 applies,except as follows.
Addition:
*SpO2 accuracy of pulse oximeter equipment
50.101.1 *Specification
The SpO2 accuracy of pulse oximeter equipment shall be a root-mean-square difference of less than or equal to 4,0% SpO2 over the range of 70%to 100%SaO2.

The declared ranges of SpO2 and SpO2 accruacy over those ranges shall be disclosed in the instructions for use. The SpO2 accuracy shall be stated over the range 70%to 100%(see 50.101.2.1). SpO2 accuracy information shall be accompanied by a note reminding the reader that,because pulse oximeter equipment measurements are statistically distributed,only about two-thirds of pulse oximeter equipment measurements can be expected to fall within±Arms of the value measured by a CO-oximeter .When a pulse oximeter montitor is suitable for use with a variety of pulse oxlmeter probes, SpO2 accuracy information shall be made available for each type

50的准确性,营运数据
的IEC 60601-1:1988 ,第50条适用,除如下。
此外:
*血氧饱和度的准确性,脉冲oximeter设备
50.101.1 *规格
该血氧饱和度的准确性,脉冲oximeter设备应一的均方根差异小于或等于4,0 % ,血氧饱和度超过范围的70 %至100 % ,血氧饱和度。

申报的范围,血氧饱和度和血氧饱和度精度超过这些范围,应予以披露,在使用指示。该血氧饱和度的准确性,应予以载明的范围超过70 %至100 % (见50.101.2.1 ) 。血氧饱和度的准确性,资料应附有一份说明,提醒读者,由于脉冲oximeter设备测量统计分布,只有约三分之二的脉冲oximeter设备测量可以预料,属于±武器的价值来衡量一个合作, oximeter当一个脉冲oximeter montitor适合用在各种脉冲oxlmeter探头,血氧饱和度信息的准确性,应提供每类脉冲oximeter探针。额外的血氧饱和度的准确性,规格超过其他范围也可提供。

如果血氧饱和度的准确性,要求在一个范围内低于65 % ,血氧饱和度是,血氧饱和度的准确性,应在一额外的范围跨度饱和度不超过20 % ,血氧饱和度。

例如一指定的血氧饱和度的精度范围内的60 %至80 % ,血氧饱和度。
示例2指定的血氧饱和度的精度范围内的60 %至70 % ,血氧饱和度..

检查遵守下列规定50.101.2和检查所附的文件。